What ionization method will be tried out first when one wants to do LC/MS analysis? Will it based on molecule polarity? Thanks.
-
- Posts: 34
- Joined: Mon Aug 30, 2004 1:46 pm
Hi Jan,
I hardly strongly recomend you to see what's on structure: If there is a N in the structure and it's not to big (high molecular weight) I would begin with ESI in positive mode.
But it also depend in the HPLC mobil phase that you are using because some compounds could produce a bad spray with ESI........
More information will be usefull to help you.
Bye....
I hardly strongly recomend you to see what's on structure: If there is a N in the structure and it's not to big (high molecular weight) I would begin with ESI in positive mode.
But it also depend in the HPLC mobil phase that you are using because some compounds could produce a bad spray with ESI........
More information will be usefull to help you.
Bye....
-
- Posts: 1073
- Joined: Mon Aug 30, 2004 5:42 pm
- Location: Atlanta, GA, USA
Hi Jan,
I started by writing a long reply of all the factors that you need to consider but that is not what you asked. So...
Personally I start almost always with ESI except if I see that the molecule seems quite hydrophobic and without any chargable functional groups. If you have a molecule which response is the same with both ionisation techniques then I would go with APCI as it has been proven that you have less ion suppression from matrix interferences...
Kostas
I started by writing a long reply of all the factors that you need to consider but that is not what you asked. So...
Personally I start almost always with ESI except if I see that the molecule seems quite hydrophobic and without any chargable functional groups. If you have a molecule which response is the same with both ionisation techniques then I would go with APCI as it has been proven that you have less ion suppression from matrix interferences...
Kostas
-
- Posts: 34
- Joined: Mon Aug 30, 2004 1:46 pm
Ops a big mistake !!
It should said: I strongly recomend you to study the structure to decide the kind of ionization that you are going to use.....
I agree with Kostas reply....
I hardly recomend you to see what's on structure
It should said: I strongly recomend you to study the structure to decide the kind of ionization that you are going to use.....
I agree with Kostas reply....
- tom jupille
- Site Admin
-
- Posts: 4978
- Joined: Wed Aug 11, 2004 4:55 pm
- Location: Alamo, CA / USA
To Basil: the new Forum software allows you to edit or delete your own posts (to correct for typos!). Look for buttons in the upper right of the message box.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
5 posts
Page 1 of 1
Who is online
In total there is 1 user online :: 0 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Most users ever online was 1117 on Mon Jan 31, 2022 2:50 pm
Users browsing this forum: No registered users and 1 guest
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science