split flow – 35.7 ml/min, total flow 39.3 ml/min. The tech checked for leaks through out the entire system
The general.m method that we use for the analysis is a bit different being in a splitless mode, start temp of 60C, Initial time 2 mins, rate 30C, 300C for 5 mins.. Front Inlet splitless 250C, 5.30 pressure, purge flow 35.7, total flow 39.3.
We created a new method called gensplit.m where the only thing we changed in the method was the split mode and the peaks were greatly reduced.
The questions, other than the problem stated above, are:
1. Why are we seeing these peaks? What could they be – water, nitrogen, oxygen??
2. Is it really a problem to run an autotune using a different method than what you are running the samples on? I don't think so as the autotune is checking the MS of the instrument. When I compared the startup.m to the general.m the only differences are in the GC parameters and none in the MS parameters.
3. Any advantages or disadvantages to running split or splitless?
Some people think that is moisture built up in the GC system and when we do the startup.m, the moisture is burned off as the peaks are lower if the autotune is done immediately after the startup.m methods ends. But could it be another contaminant?
The analysis on this GC/MS is for controlled substances, qualitative analysis only.
Thank you for your help.
Christy