calibration curve, chemstation

Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.

3 posts Page 1 of 1
I was wondering if anyone knew how to set up the calibration curve so I can have Area standard/Area ISTD as one axis and Amount standard/Amount ISTD as the other. I can't seem to find a way to set it up to do this. My other option is using a spreadsheet but I thought I would try this first.

Thanks

Andrew

Its not something that I'm althogther familiar with...new to ChemStation myself. But the first thing I'd ask is if you have selected the option for Internal Standards within the calibration settings?

I reckon this would be your first port of call.

The above reply from Anthony is correct. If you have a multilevel calibration table set up, select in the report style that you want to have an internal standard report. If you then look into the calibration table and look at the the calibration curve, you will see that the axis have changed to Area Ratio and Amount Ratio.

Gasman
3 posts Page 1 of 1

Who is online

In total there are 407 users online :: 1 registered, 0 hidden and 406 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google Adsense [Bot] and 406 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry