How should Splitter Flow be adjusated

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
in online GC application, How should splitter flow should be adjusted. When injection Valve is 'on' to delute the sample or otherwise when sample is being flushed
Regards,

Online GCs usually have specialist inlet plumbing that does not necessarily conform to the usual split-splitless inlet of a bench GC. You need to refer to the operating instructions for the specific instrument that you are working with.

Peter
Peter Apps
Hellow

I want to know the mean of 100:1 split ratio.

how it is releated to total flow, split vent flow and column flow

praveenpaliwal,

If your column flow is 0.5 mL/min, your are also sending 50 mL/min out the split vent and only loading 1/100th of the sample injected on column. If your column flow is 2 mL/min you are sending 200 mL/min out the split vent and only loading 1/100th of the sample on the column. In the first case you might be supplying a total flow to the injector of, perhaps, 55 mL/min: 0.5 for column, 50 for the split and 4.5 for the septum purge. Rarely works this cleanly but you get the idea.

Best regards.
4 posts Page 1 of 1

Who is online

In total there are 16 users online :: 1 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: CharlesPar and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry