Detector linearity calibration
Based on recent information during an LC/GC webinar on AIQ, the presenter mentioned that you should have "global" type procedures for calibration so all instruments (HPLC) are being calibrated in a similar manner.
My question pertains to detector linearity. How does your company perform detector linearity, and does the procedure account for the different absorbance ranges of the different manufacturer models?
Does anyone simply use the ASTM procedure by running 5 standards, then calculating response factor? (Requirements being the rsd of the response factors must be less than 5%) Or does anyone simply calculate slope and intercept?
I'm just curious if there is an industry standard. Thanks.
Russell
Dionex uses 5 certified caffeine standards then calculates slope, intercept, r^2 and RSD of the linear fit. The specifications vary depending on the model, but for the UVD3400 they are r^2 > 0.9997 and RSD < 3.0%.
This test pre-requires that the detector noise and drift, the pump flow rate, and the injector repeatability all be within specs.
Mark Tracy
Senior Chemist
Dionex Corp.
In total there are 29 users online :: 2 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Users browsing this forum: Bing [Bot], Google [Bot] and 27 guests