How many individuals are doing individual weigh out for linearity
and how many are doing dilutions from a stock solution?
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
For assay analysis is max. absorbance 0.5AU suitable but when you analyze impurity profile at 200-210nm with LOD 0,05 area %, you must have major peak about 1-1,5 AU then its linearity must be confirmed ( or use diluted reference as second run). Because when is its response nonlinear you get incorrect result. I observed nonlinearity for some compounds. Mainly near 1,5 AU but in one case the compound behave nonlinear even about 0,9 AU (I supposed it was evocated by its fluorescence). Instrumentation Agilent 1100 with VWD detector.I am going to sound heretical here. I think linearity testing is a waste of time! With the proviso that the max absorbance is 0.5AU and the instrument is properly maintained I cannot see the value of testing the obvious.
Does anyone have any examples of non linearity in these circumstances?
Happens to me a few times every year, although my methods linear ranges can go up to absorbance of 1, with instrument detectors claimed to be linear to >2, eg Agilent DAD and VWD.With the proviso that the max absorbance is 0.5AU and the instrument is properly maintained I cannot see the value of testing the obvious.
Does anyone have any examples of non linearity in these circumstances?
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.