- Posts: 392
- Joined: Thu Sep 08, 2005 4:12 am
I have the following conditions,
Column TSK-GEL Phenyl 5PW RP, 4.6x100x10um,
MP ACN:H2O=20:80 to 55:45 in 25 min Flow 1ml/min
wavelength 230nm
When I have consistently seen one contamination peak at 8.5 min out of no injection(0 ul) or 1ul of (H2O/ACN). This peak has the same retention time as the analyte I am testing so it interferes my test.
LC gave a perfect base line when column was removed off the LC.
After I wash 1) 90% ACN overnight, and 2) 1ML 0.1M NaoH separately, this column gave the same peak in both cases; I tried 4 different of DI water, the same peak there. The peak was present in 30 consecutive injection of 0 ul .
But when I repeated the above condition with a TSK GEL Octadecyl 4PW column 150x4.6x7um, my analyte eluted at 13 min. The contamination peak was also seen at 13 min in the first 3 injections but this peak disappeared in the following 20 injections.
The contamination peak area is about 2k to 5k. Yes, the peak became bigger when the equilibration time was longer.
Question 1, why the two column gave different phenomenon with the same mobile phase? And do you have suggestion to avoid the contamination peak with the 100mm column?














 
																							
 
 