reconditioning a column
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
			2 posts
							Page 1 of 1
					
	
						we have a formulation that contains keltrol as a thickener. After several injections onto a column the peaks start to split. Our samples are clear before injection, but we suspect the keltrol as being the cause of the split peaks. Our mobile phase contains an ion pair. We have reverse washed the column with water at 50 degrees followed by phase but we still get split peaks. - does anyone know how we can remove the keltrol and clean the column up. desperate for suggestions.
					
									- 
								
 
- Posts: 1233
- Joined: Tue Oct 11, 2005 7:20 am
						Perhaps some details might help - such as manuafcturer, column type, size, mobile phase composition ( isocratic or gradient - if gradient your column's probably doomed ), ion-pair agent and concentration, and flow, sample prepartion and solvent, and injection volume.
Have you got a guard column in place?. If so, what happens when you replace it - does the pressure drop?. Did you monitor the pressure when you were injecting the samples?.
I'd expect the Xanthan gum to form a goo over the column if you tried a gradient, and the backpressure would climb rapidly, possibly forming a void in front of the column. If that's the case, it's unlikely to go away.
You need to test the column with the manufacturer's test mix, and compare resolution, peak shape and backpressure with the original certificate. Make sure you wash all the ion-pair agent from the column first, and carefully monitor pressures. Slow downward trends in pressure when reverse flushing may suggest you will be able to wash the gum off with persistence and warm water. If peak shapes are OK, it's worth persisting with flushing.
If it's a void ( test solution shows all peaks split, and backpressure is similar to CoA, not really high ) the best solution would be to purchase a new column, however the second best solution may be to take the column to a pressure about twice normal operating pressure using a high flow rate for about 10 minutes, gradually decrease flow to zero. Obviously don't exceed column's maximun specified pressure.
Carefully unscrew front fitting - if there's a void, carefully fill with material from the back end of a used guard column, or some other source of similar material. Repeat the pressurisation process, and refill if necessary. I've had columns go from 40% of orginal plates to 85% with this nethod, and got another six months of life.
Now, address the cause - change your sample preparation. You need to either trap the Xanthan on a resin, or perform a sample purification, possibly using SPE or Liquid-liquid extraction of your analyte into an organic solvent that the gum doesn't like, with teh solvent choice depending on your analyte. Also, ensure you have aguard column in place, they usually are a little cheaper to replace.
Please keep having fun,
Bruce Hamilton
									Have you got a guard column in place?. If so, what happens when you replace it - does the pressure drop?. Did you monitor the pressure when you were injecting the samples?.
I'd expect the Xanthan gum to form a goo over the column if you tried a gradient, and the backpressure would climb rapidly, possibly forming a void in front of the column. If that's the case, it's unlikely to go away.
You need to test the column with the manufacturer's test mix, and compare resolution, peak shape and backpressure with the original certificate. Make sure you wash all the ion-pair agent from the column first, and carefully monitor pressures. Slow downward trends in pressure when reverse flushing may suggest you will be able to wash the gum off with persistence and warm water. If peak shapes are OK, it's worth persisting with flushing.
If it's a void ( test solution shows all peaks split, and backpressure is similar to CoA, not really high ) the best solution would be to purchase a new column, however the second best solution may be to take the column to a pressure about twice normal operating pressure using a high flow rate for about 10 minutes, gradually decrease flow to zero. Obviously don't exceed column's maximun specified pressure.
Carefully unscrew front fitting - if there's a void, carefully fill with material from the back end of a used guard column, or some other source of similar material. Repeat the pressurisation process, and refill if necessary. I've had columns go from 40% of orginal plates to 85% with this nethod, and got another six months of life.
Now, address the cause - change your sample preparation. You need to either trap the Xanthan on a resin, or perform a sample purification, possibly using SPE or Liquid-liquid extraction of your analyte into an organic solvent that the gum doesn't like, with teh solvent choice depending on your analyte. Also, ensure you have aguard column in place, they usually are a little cheaper to replace.
Please keep having fun,
Bruce Hamilton
			2 posts
							Page 1 of 1
					
	Who is online
				In total there are 31 users online :: 0 registered, 0 hidden and 31 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: No registered users and 31 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 31 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science














 
																							