 ,I have a recurring but intermittent problem. We are currently analysing dissolution samples by HPLC and every now and then we get a higher than expected result for one of the timepoints, which at the next timepoint returns to 'normal' (i.e 10 mins=85% dissolved, 15 mins=109%, 30 mins=100%). Upon re-injection of these vials, the 109% would be 99% for example - fitting back in with the expected profile. On closer analysis of the particular chromatograms we usually notice that the solvent front peak is also bigger. So far we have just documented a 'suspected HPLC injection error' and reported the repeat injection result, but it's happened a few too many times now to ignore. The HPLC system is in a calibrated state and system suitability is within acceptance limits both pre- and throughout the run. The injection volume is 100 µl (which is the maximum calibrated volume on the system). The phenomenon has happened on more than one HPLC system in our lab.
Any suggestions?














 
																							