hi:
I meet a strange phenomenon!have you ever met before?
we have a 50mmID*250mmL column here,when we test the column efficency with
mobile phase: 50%ACN:50water
Detector was 254nm,
flowrate:55ml/min
silica:C4
testing probe:paraben
everything is ok!but when I change the mobile phase to 50%MeOH:50water,the column efficency will decrease a lot ,and the T-factor will encreased a lot !
I just want to know why is so ?and do you have some recommendation on testing the column efficiency(50mmID*250mmL,C4)
thanks !

Methanol and acetonitrile are not the same. Changing from one to the other can dramatically affect selectivity, retention, and peak shape depending on the compound. You calculate column efficiency using the Van Deemter equation. Test column efficiency under the conditions that you plan to use for your analysis.

Hi,
To substitute %50 ACN, I would suggest %62.5 MeOH to obtain similar retention.
As a general rule, you can substitute ACN with MeOH but with proper ratios. 50% MeOH has a roughly equal solvent strength with 40% ACN.
But do not forget that this is not the only thing, Sassman already explained selectivity and peak shape issues.
For testing column efficiency: a column will have different efficiencies for different methods. There may be a lot of ways, virtually all of them previously discussed at the forum.
If you have a good correlation of paraben efficiency and your effective column life (for your method), then stick to your paraben injecting method.
Good luck,
Bulent

The viscosity of the mobile phase has a big effect on the efficiency of the separation. 50% MeOH is about three times as viscous as 50% MeCN.
Mark Tracy
Senior Chemist
Dionex Corp.

Agree with above posters. You'll also get more reproducible results with parabens if you add a little acidity to the mobile phase (e.g. acetic or phosphoric).
5 posts Page 1 of 1

Who is online

In total there are 21 users online :: 1 registered, 0 hidden and 20 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 20 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry