						Sunjay,
There are three different ways to measure noise in the chromatogram (4 if you count standard deviation):
1) 'peak-to-peak' noise (Np-p)
2) 'peak-to-valley' noise (Np) (This is also called "peak noise").
3) RMS (root mean square) noise
An electronic system such as the Agilent CDS is most likely measuring RMS noise as that is an easy to do electrical measurement.
I don't remember how the ASTM method works.
Np-p = 2 X Np
and
N(RMS) = ~0.7 Np
As Mark noted, you need to know how the noise is being determined. That way you can see why there are differences. Check with your vendor.
The best guidance that I can think of is ICH Q2(R1). Please see this link:
http://www.ich.org/LOB/media/MEDIA417.pdf
I have usually used Np-p to determine the S/N, but, on occasion, I have used the standard deviation. It all depends on your SOP or validation protocol. The important thing is to document the technique for the noise measurement in the validation protocol and report and in the method.
Regards,
Dan