I'm using a XBridge C18, 5µm, 19x100mm and get a backpressure of about 1700 psi @15 ml/min if I had it correct in mind.
So the pressure for a 19x150 mm I would expect is about 2500 psi @15ml/min.
Maybe there's a possibility to change the pressure limit so you can set a higher one. Even with this, I think you won't be able to run your colum @30ml/min.
But keep in mind, that you'll lose some performance at flow rates too high. 15 ml/min @19mm will turn to about 0.9ml/min @4.6mm column, which is near the optimum of the VanDeemter curve (for 5µm particles).
Maybe this could turn into a higher loadability (:?:

)