Dear Friends,

I would like to hear from anyone that has experience working separations at low pH (2-3) on Polaris A columns. No clear information is provided by the manufacturer as to the nature of the polar embedded phase, or its stability to buffers and low pH conditions.

I have had some bad experiences with similar columns containing "epoxide" phases, and before exposing the Polaris A columns to "harsh" conditions I would like to know more about their stability.

Please remember that ther are two types of Polaris columns, Polaris A and Polaris-ether. I am interested on the A type.

Thanks,

josebenjamin

josebenjamin

Polaris-A columns use an amide linker before the C18. It should be as robust as any other C18 column. According the the literature it is stable don to pH 1.5. I use one with ammonium acetate buffer and routinely get 700-1000 injections of reasonably "dirty" samples. I discard columns when I start to get peak splitting most likely due to accumulation of protein at the head of the column.
2 posts Page 1 of 1

Who is online

In total there are 28 users online :: 1 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry