lmb » Thu May 24, 2007 6:24 pm
Dear wrighty,
To keep the same retention times and resolutions of all peak pairs after trimmin a column or replacing it with a new one (hopefully from the same manufacturer), you do not need to worry about changes in a column length or average gas velocity. What you need to do is to make sure that hold-up time, tM, (also known as dead time, void time, etc.) is always the same.
There are many way to restore tM after making the changes. For example, you can do this:
1. Choose any good peak in a chromatogram and use it as a reference (it might be a slightly retained peak like methane or butane, but it does not have to be slightly retained one).
2. Keep a record of retention time (RT) of that reference peak. (If you did not do it before trimming or replacing a column, try to find that RT from previous chromatograms of the same sample).
3. If, after trimming or replacing a column, RT of the reference peak changed, adjust column pressure (or flow) to restore that RT to its previous value. You might need to do several iterations because there is no good general formula for pressure correction in all cases. The only certain thing is that the pressure change should always be in the same direction with RT change (for example, if RT became shorter, reduce the pressure). Although there is no simple formula for all cases, there are recommendations for special cases.
3a. If analysis is isothermal or if the reference peak is slightly retained, %-change in pressure across the column should be the same as %-change in RT (whether a column was trimmed or replaced).
3b. If a column was trimmed, measure the length, of the peace that was cut, find its approximate relative value in relation to approximate column length. Relative reduction in pressure across the column should be twice the relative reduction in column length. This works for isothermal and temperature-programmed conditions.
3c. The most uncertain case is replacing a column in a temperature-programmed analysis and the reference peak is not slightly retained. Try to make a relative change in pressure across the column that is five times larger than relative change in RT. This might or might not reduce the number of iterations to restore RT.
Notice that, in all special cases, you deal not with pressure reading on GC instrument (gauge pressure), but with the pressure across the column. For example, in case of GC-MS (vacuum at column outlet) pressure across the column is gauge pressure + 1 atm.
Finally, as far as I know, Agilent sells RTL (retention time locking) Software essentially based on theses procedures.
Good luck,