precolumn to use for separation of H4SI(monosilane) from air

Posted: Mon May 07, 2007 3:16 am
by tigerk2001
Dear all
I want to measure H2-O2-N2-CH4-CO-CO2 in pure H4SI(monosilane) and I would like to know which column(or configuration) is the best to use as pre column for the separation of the air from H4SI.
The second column can be a Molecular sieve 5A for H2-O2-N2-CH4-CO and a porapak for CO2. A ShinCarbon can also be used for all permanent gases and CO2.
My biggest concern is for the separation of H4SI from the air.
thanks

Posted: Mon May 07, 2007 1:25 pm
by chromatographer1
A Carboxen 1000 should separate the silane from air.

best wishes,

Rodney George
Senior Research and Development Scientist
Gas Separations Research
Supelco
595 North Harrison Road
Bellefonte, PA 16823

814-359-5737 voice
814-359-5459 fax
rodney.george@sial.com

Posted: Mon May 07, 2007 4:38 pm
by chromatographer1
Dear Tiger,

Based on bp, silane should elute betwen methane and ethane.

If you used a long high loading methyl silicone column or a moderate length Carbopack B column you should be able to separate silane from air and CO readily.

best wishes,

Rod