Mobile phase:17 g of monobasic ammonium phosphate per L, adjust with phosphoric acid to a pH of 3.0, and mix.
After several times, we are getting very early retention time.
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
Thank you very much!Your column could lose capacity for two reasons: the sites are clogged with polyvalent ions or the bonded phase is coming off. If it is the second, there is no hope. If it is the first, you can regenerate with a high ionic strength wash such as 1M ammonium sulfate or a chelating agent like pH 4 citrate.
Silica-based ion exchange columns are not as tough as typical C18 columns, and much less resilient that polymer-based ion exchangers. Treat your column gently and follow the manufacturer's recommended use and storage conditions.