Hi, everyone!
I'm trying to assay oleic acid and monoolein in a farmaceutical product using RP chromatography with ELS detection. I'm not very familiar with this detector and when I found out that the precision was poor I supposed it could be due to detector fluctuations. I added palmitic acid (eluting between the analytes of interest) but its area remains constant while the others vary some between injections.
I would very much like some suggestions of how to increase the precision and what to think of when doing method development on an ELSD.

Thanks

Check your ELSD temperature. These carboxylic acids can parially evaporate in ELSD thus reducing amount of acid you see. Try to drop temperature to 35*C in ELSD.
We had this problem when we developed a method for C12, C14 and C16 carboxylic acids.

regards,

Vlad
2 posts Page 1 of 1

Who is online

In total there are 29 users online :: 2 registered, 0 hidden and 27 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 27 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry