My setup for the determination of low concentrations of water in solvents is:
- DANI 85.10 with macro TCD (~ 10 ml/min flow required) and 651.1 electrometer
- Carrier He 12 ml/min (5.0 purified with molsieve and SUPELCO high capacity purifier)
- Porapak R 100/120 1/8" SSx6 ft.
- Injector: 210 °C
- Detector: 280 mAx1; x10; Cell: 220°C
- Oven: 215 °C isothermal
- Peak of water at around 1 min.
- Injection volume: 10 yl
It even allows for the detection of water in ethanol (both peaks are nicely separated).
The determination of water in a solid matrix is a classical application for headspace GC.
Hope this helps you