Dear all,

The USP and Ph.Eur method for 5-aminosalicylic acid (5-ASA) is a ion-pair, reversed phase method (C18 column). I wanted to analyse the substance with LC-MS so I tried a Primesep 100 column, and got good retention with a simple mobile phase (0.05% TFA in 30% ACN).

The thing is that I now see a "gigantic" impurity (5-20 area%) eluting a couple of minutes before the main peak. The impurity is present in all tested 5-ASA batches, including the newly purchased CRS standard from USP. The LC-MS and UV data suggests that the impurity is related to 5-ASA. Mass (M+H): 304 (154 for 5-ASA), UV max (three) shifted towards longer wavelengths. The mass is not mentioned in any of the references I have found about 5-ASA degradation.

The peak is not present in the blank injections, or in the injections of the known impurities.

I have isolated the peak, but it is not stable after isolation (t½ about 30 min). And it is converting back to 5-ASA! I have tested to dissolve the substance in many solvents/buffers: no difference

Right now I don't know what to do with this information. Is it a true impurity, or could it be some kind of artefact still? Some kind of unknown equilibrium in the aqueous phase?? If it would be a "real" impurity it should make the original method unprecise, but I never see any problem there. Any good suggestions?

Mattias, when you say it's converting back to 5-ASA, how are you measuring that? By the same method you used to generate it?

And does it return to that same ratio of "impurity" to 5-ASA?

My thought on first reading was a dimer or aggregate, but that doesn;t explain the 3 unit difference, so I'm out of ideas.

You might also try cross-posting this in the LC-MS section and see if some of the more MS-oriented people might have some suggestions.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374

Hi Mattias,

We run this method with various mobile phases (UV and LC/MS):

http://www.sielc.com/compound_286.html

I need to look at individual injections and see if we observed something similar.
I would suggest reproducing method from the link using formic acid instead of TFA. May be TFA catalyzes some kind of condensation/decarboxylation.
There is nothing unusual to see some additional impurities with Primesep columns. We once analyzed “pure standardâ€

Tom> I have collected the peak manually (LC-vial in the outlet of the detector). Then I reinjected it in the same system. First injection gives only one peak ("impurity"), then two peaks (impurity +5-ASA), at last only 5-ASA.

Vlad> Thanks for the link! I will try formic acid and see how it looks. The detection is at about 300 nm, so there should be no problem.

I'll let you know how it goes.
4 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry