regarding less sensitivity of LC-MS-TOF

Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

2 posts Page 1 of 1
which parameters we need to change and up to what extent (flow of drygas, nebulizer,potential of quadrupole and hexapole, collision energy) in LC-MS-TOF to increase the response of the drug e.g. hydrochlorothiazide
All of them!

MS intensity response is quite sample specific, all the parameters you list can make a difference.

However:
I would start with the ESI needle capillary V, probe position and cone or source V settings.
Then I would adjust the gas flow
Finally I would adjust the collision energy.

I would not adjust the hex or Quad settings, as these tend to be more instrument specific than sample specific.

All parameters can be adjusted by +/- 50% of their current value. be aware that some parameters interact eg porbe position and source V.
2 posts Page 1 of 1

Who is online

In total there are 42 users online :: 0 registered, 0 hidden and 42 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: No registered users and 42 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry