						Why do I see adducts of +46 in Negative electrospray and how can I minimize it?  I use water/ACN mobile phase with 0.1% formic acid on a ZQ single quad.
					
									- 
								
 
- Posts: 9
- Joined: Thu Dec 22, 2005 3:53 pm
						Possibly these are formic acid (from the mobile phase) adducts?  Formic acid has a M.W. of 46.
Almost anything you do to reduce adduct formation may impact your target analyte sensitivity. But normally you try increasing the source temperatures (desolvation/source block) and gas flows (desolvation/cone) as well as your cone voltage (this last one will almost certainly affect your target analtye). I've been told that adjusting your probe position can also help.
If it's the formic acid giving the adducts, reducing the mobile phase flow rate into the source (splitting) or decreasing the % of formic in your mobile phase might be worth a try too.
									Almost anything you do to reduce adduct formation may impact your target analyte sensitivity. But normally you try increasing the source temperatures (desolvation/source block) and gas flows (desolvation/cone) as well as your cone voltage (this last one will almost certainly affect your target analtye). I've been told that adjusting your probe position can also help.
If it's the formic acid giving the adducts, reducing the mobile phase flow rate into the source (splitting) or decreasing the % of formic in your mobile phase might be worth a try too.
- 
								
 
- Posts: 216
- Joined: Mon Apr 04, 2005 12:26 am
						don't know why you see M+46 ions with formic.  However, not unusual to see M+45 ions when using formic or M+59 ions when using acetic in electrospray.
Especially for things that don't easily form negative ions. For example, PPG 1000 forms very significant acid adducts ions in negative ion electrospray. We use them plus the acid dimer-H ion (m/z 119 for acetate case) to calibrate in negative ion electrospray.
									Especially for things that don't easily form negative ions. For example, PPG 1000 forms very significant acid adducts ions in negative ion electrospray. We use them plus the acid dimer-H ion (m/z 119 for acetate case) to calibrate in negative ion electrospray.
Sailor
				
			3 posts
							Page 1 of 1
					
	Who is online
				In total there are 24 users online :: 2 registered, 0 hidden and 22 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
 
		
Users browsing this forum: Ahrefs [Bot], Google Adsense [Bot] and 22 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Ahrefs [Bot], Google Adsense [Bot] and 22 guests
Latest Blog Posts from Separation Science
    Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
    
    
    Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
    
    
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science














 
																							