- Posts: 7
- Joined: Fri Feb 23, 2007 5:51 pm
I am trying to collect some calibration data for the GC, and I have been having a hard time getting rid of the residues from my previous analyses. For example, I inject 1 ul of 10% acetic acid in water, and do a blank run right after. I do see NO peak in my blank run(s). Then I inject distilled water. (I use FID so I shouldn't see any peak.) And there it is! My acetic acid peak... Its area is 1/10 of its area in my original analysis. If I continue injecting distilled water its area decreases gradually. Other compounds that I injected also showed the same result.
Here is the information on the system I am using:
GC: HP 5890II
Column: AT-Q
Column flow: 4 cc/min
Aux flow: 26 cc/min
H2 flow: 30 cc/min
Air flow: 400 cc/min
Total flow: 100 cc/min.
Septum purge: 3 cc/min.
Oven temp: 40 to 180 at 15 C/min
FID: 200 C
Inj: 200 C
GC equipped with Electronic Pressure Control system
I am using an HP autosampler to inject my samples and I tried having it wash the syringe 30 times with my solvent (water), 15 times with my sample (again water), and pump 15 times with my sample. I am pretty sure this is not an issue with the syringe. I am thinking my sample is getting trapped somewhere in my injection port and with my next sample being swept into the column. But I don't know for sure. Has anyone ever come across with such a situation? By the way, there is an on-line valve installed on my carrier gas line.
Thanks,
Hakan