Dear friends,
I am doing the LCMSMS analysis of two drug molecule (analyte & metabolite) in a single run. I am using API 2000, ESI source,MP of ACN : 0.1% Formic acid , chromolith C18 performance column, run time 2.5 min . The extraction process is by LLE by diethyl ether.
the problem I face is that the linearity is totally out since sensitivity of the analyle is reduced to almost 25% at higher CC conc CC7 CC8.,when compared to CC1 & CC2. whereas the metabolite is coming perfect.

One more information is that my machine was calibrated few days ago and earlier to that i had got the same method with good linearity.

Please send me your valuable suggestions at the earliest possibe....very urgent.

Shan

Try to find whether the linearity problem is due to your MS or to problems with the extraction. If you inject your analyte as standards with comparable conc. to your extracted plasma samples, is that linear?

regards Bert
2 posts Page 1 of 1

Who is online

In total there are 14 users online :: 2 registered, 0 hidden and 12 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Google [Bot] and 12 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry