acetate and lactate by IC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
I've been trying to separate aceteate and lactate by IEC using a Dionex AS11 column and a dilute (< 1 mM) NaOH eluant with conductivity detection. The peaks are poorly resolved and overlapping at higher concentrations. The column literature suggests that this separation should be possible.

Does anyone have any suggestions?

Is yours an AS11 or an AS11-HC? The AS11-HC is much better for this application. Have you looked at the product datasheet?
http://www1.dionex.com/en-us/webdocs/41 ... HC_V15.pdf
It compares the two columns in Figure 1. Also, try raising the temperature to 30 °C.

You can also consider the just-introduced Acclaim Mixed-Mode WAX. The user's guide is at:
http://www1.dionex.com/en-us/webdocs/48 ... olumns.pdf
Mark Tracy
Senior Chemist
Dionex Corp.
2 posts Page 1 of 1

Who is online

In total there are 41 users online :: 1 registered, 0 hidden and 40 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 40 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry