Molecular sieve 5A regeneration

Discussions about GC and other "gas phase" separation techniques.

4 posts Page 1 of 1
Hi,
Please explain me what is the best way to regenerate a molecular sieve 5A. I normally use Helium or Argon as carrier gas.

I problem I met is that I lost the O2 peak in low ppb concentration. Is there something special to do in the column regeneration process.

thank you

If your peak has suddenly disappeared, check to make certain that your detector still has good sensistivity for other trace gases, and the correct sampling loop size with flows working OK.

Conditioning depends on the column maximum operating temperature, and whether any other columns are in the oven. For PLOT fused silcia columns, they usually have a maximum of about 190C.

I prefer to condition a plot column overnight at 190C with helium, but the really important aspect is to make sure that you maintain at least a low flow continuously after conditioning. If you have a packed column in stainless steel you can go up to 265C, but you should follow the manufacturer's recommendations.

The other aspect is to ensure that your carrier gas purifiers are all functioning correctly, and the carrier gas is pure. Make certain all your detector makeup and reference gases are also pure.

Bruce Hamilton

I have a home made GC I use for packaging gas two parallel columns Haysept the other MS5a the columns go to different sides of a TCD detector pair. Responses go up and below base line.
When MS5a column is getting near regeneration the oxygen and nitrogen peaks shift towards the injection point the dont decrease in size.
I suspect that you are looking at a detector problem or a loss of sample at the injection point.

Bob M.
Remember by just being alive you risk dying.
The best thing to do if you lose your oxygen peak (which is a common problem actually) is to use air or oxygen to activate your sieve.

Activate slowly up to 250 °C and hold for 4 to 24 hours (experiment to find out how much activation you require).

0.125 in columns use 20-50 mL per min. 0.0625 inch column use 5 to 20 mL per min.

Rodney George
Senior Research and Development Scientist
Supelco
4 posts Page 1 of 1

Who is online

In total there are 46 users online :: 1 registered, 0 hidden and 45 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 45 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry