- Posts: 9
- Joined: Fri Mar 02, 2007 3:55 pm
I am fairly new HPLC and very new Ion Exchange HPLC. I am attempting to develop an IEX method for a protein (about 64kD) at a pI of 4.0. I first went with a weak anion exchange phenomenex column (Shodex IEC DEAE-825 8mmx75mm) and for the mobile phase I was using 20mM piperazine pH 5.0 and buffer b was the same thing plus 1M NaCl. For the gradient I started with going from 100% A to 100% B for 60 minutes at a flow rate of 0.5mL/min. And I got a peak at 3 minutes. I have tried changing a few things such as going to a strong anion exchange (also by phenomenex IEC QA-825 same dimensions) and tried going down to 5mM piperazine, and the peak moved alittle to about 5 minutes I also tried bis-tris and that didnt work either. The bottom line is that my boss is not comfortable with an assay where the peak comes off in 5 minutes, he would prefer to see it come off at about 10-15. Is this a common retention time for ion exchange? Is there something else I can do increase the retention time? Thanks in advance for your help!