Reference Wavelength

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

3 posts Page 1 of 1
What is the function or significance of reference wavelength in PDA detector?

A www search should fully answer your question, as should the instriument manual.

Generally, the reference wavelength is set to a part of the spectrum that analyte peaks don't absorb, and will thus provide a correction for ambient temperature and gradient refractive index effects on the mobile phase absorption.

If you want to be really clever, you can use the reference wavelength feature to quantify an analyte in the presence of co-eluting peaks using peak suppression, however some loss of sensitivity usually occurs.

Bruce Hamilton

There has been at least one discussion on this here. Search the archives
Mark Tracy
Senior Chemist
Dionex Corp.
3 posts Page 1 of 1

Who is online

In total there are 30 users online :: 1 registered, 0 hidden and 29 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google Adsense [Bot] and 29 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry