I am having an issue with light alcohols (MeOH, EtOH, IPA, and to a lesser extent Acetone), where I have poor sensitivity and poor peak shape. The detector is MS, but I think the detection type is irrelevant in this instance and would rather have more input from those with more knowledge about GC.
The instruments are: a Tekmar HT3 (in dynamic mode), 6890 GC, 5973 MS. The phase in trap is Supelco "K", and desorbed at 220°C. GC Column is a 60x0.25, 1.4µ DB-624. Temp prog is 40°C for 1 minute, then 1°C/min to 60°C, 10° to 120, 15° to 240, hold at 240 for 5. Typical volatile program.
The peak shape for all components is great except for those listed above. The liner was new when I istalled the HT3 and new column. I have tried changing the liner without success (it really has no purpose because of the HT3 anyway).
I also reduced the desorbing time on the HT3 from 1 minute to 0.2 minutes without any improvement.
The samples are water, and the desired concentrations of each of my standards is from 0.05 to 50 ppm) Essentially I have everything working except the methanol, ethanol and isopropanol. They seem to be "sticking" somewhere in the system. Let me know if you need more information about the system to make a diagnosis, however I do not want to give away too much.
Thanks!
Example:

edit: added more details