should protocols include these checks? I cannot in good faith eliminate that criteria due to years of experience, but I need some hard facts to present to these misguided folks.
thanks!
Advertisement
Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.
."
I obviously can't comment on who needs training, but are you listening and responding to their reasons, or just seeking confirmation of your position?.In one case, a chromatographer made up about 6L ACN containing MP, and didn't degas it since there is a degasser on the instrument. Let me tell you what a mess that was!
So, everyone is confirming my gut feelings, and is showing me that there needs to be some more training.
(source: http://www.fda.gov/ora/science_ref/priv ... /jaoac.htm)Pump
Are flow rates accurate and reproducible?
It's too easy to get distracted with volumetric flasks, you only get one shot at the measurement. I prefer a 5 or 10 ml with 0.1 ml/divisions graduated cylinder. It's easier to do other things, and come back after 4 - 8 minutes to record the volume at the next full minute - assuming flow is 1 ml/min.Dear wanda50:
As a practical note, you need to use traceable timer (get one from VWR, or calibrate your timer with tone from NIST), and calibrated vol. flask for the measurement.
Oops, I probably shouldn't generalise so much, as different UV detectors will behave differently, and later lamps are supposed to be superior to older models.Bruce, I wonder about this lamp life depending on the wavelength used?