mAu and millivolts

Posted: Sun Feb 04, 2007 10:09 am
by pankajcops
are mAu and millivolts units interchangeable in HPLC in case of PDA detector?

Posted: Sun Feb 04, 2007 4:42 pm
by Uwe Neue
You need to check, if the detector output is 1 V per 1 AU. This may be, or may not be. Look in the manual.

Posted: Mon Feb 05, 2007 5:45 pm
by Mark Tracy
I assume from the context of the question that you have a PDA connected through some interface box. Check the configurations of each.