increasing baseline at HPLC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Im performing TAG analysis of corn oil ussing a 100% methanol gradient to 100% isopropanol gradient on an 250mm 4.6 mm diameter 5 micromeer C18 coloumn and uv detector at 210nm and got satisfactioary results but after 3 mounts I wanted to repeate my run and had an increasing base line till 2000mabsorbance could it be due to dissolved oxigen in isopropanol what should I do plz help me

Dissolved oxygen has only a very small effect at 210 nm. You most likely have contamination. First check the methanol and isopropanol to see if they are contaminated. Then try washing the column 15 minutes each with methanol, then acetone, then dichloromethane, then back to methanol. Be patient.
Mark Tracy
Senior Chemist
Dionex Corp.
2 posts Page 1 of 1

Who is online

In total there are 37 users online :: 3 registered, 0 hidden and 34 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Bing [Bot], John Guajardo and 34 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry