I have a very very approximate way of doing this analysis when a "semi-quantification" is required here, and cost and time are in priority.
Things need to be considered, such as that the expansion in the injector may not be reproducible, some will be lost via the split (splitless must be in use) and also via the septum purge. Also the actual temperatures and pressure of expansion may be different inside the injector.
For example, a liquid of toluene standard in CS2 is prepared, and 1 uL of aliquot is injected in a spitless injector. This liquid (1 uL) will be expanded into 322 uL in gas in injection port using HP Pressure/Flow Calculator. A calibration curve can be obtained for toluene with the knowm amount of toluene and the integrated area obtained from GC run. Then 500 uL of gas sample from an airbag in which the toluene is present is injected.
If the quantification of the gas sample based on the above calibration curve is given to be 100ug, for example, so a factor of 0.644 from 322uL/500uL is used to correct the raw data of 100ug. The final result should be 64.4 ug (100 x 0.644) of toluen in the sample bag. Does it sound OK?
Cheers
Jimmy