When using our Waters HPLC’s (Alliance 2690 and 996 PDA) we see extremely noisy baselines. We use Thermo's Atlas as a data acquisition system, and record the analog channel out and convert it with a Thermo A/D box. To the best of my troubleshooting, it is not a chemical noise issue, and not related to flow through the detector. The mobile phase is acetonitrile and buffer (ammonium acetate). The detection wavelength is 278nm. I have explored increasing the bandwith of the PDA detector, and it does decrease the noise in the chromatogram, but not to a sufficient level. The Hamming filter does have a dramatic effect on the noise, but the baseline has a lot of square blockish features in it.
In the chromatogram below, the green trace is from an Agilent 1100 VWD, and the blue trace is the Waters 996 PDA, and they are on the same scale. The firmware for the 996 is the most current version, but I suspect that the noise is in the signal leaving the detector. Any comments?
I just wanted to see if anyone else has experienced this before.
Cliff