Dear, GasMan and Rod
Today, I solved the problem.
Shimadzu thechnical supporter came to Lab and repaired GC.
The problem was an air. Air was trapped by the silica gel.
Silica gel was old. Then, air was wet condition.
The time of drop of peak was the back pressure of air pump.
We changed to new silica gel. The baseline was straight in now.
But, peak dropped a little when operate pump.
(GC setting-atteuation is very high and also range is narrow; analyze trace amount of ethylene)
He recommened the use of hydrocarbon trap and inter-valve within air pump.
Thank for your sincere advice .