I am a new user to ELSD so I am researching my options (what I am trying to say is pardon my ignorance!).
I took my data and used the polynomial and log/log curves to calculate purity (obviously can't use straight area percent due to the lack of linearity). Even though the plots gave fairly similar correlations, the calculated purities were dramatically different (78% by log/log, 87% by polynomial, 94% by area normalization assuming linear response (which is non-linear)). In the log/log plot, more weight is given to the low on-column load samples in the line which led to a more dramatic increase in the calculated amount of each impurity.
My thoughts are that this is closer to the true value than using the polynomial plot. When I just look at the response factor at each level I see that my lowest point on my curve is only 8% of the response factor of my nominal load. Therefore it makes sense that there is such an extreme increase in the calculated amounts as compared to just area normalization.
The good news is that we are also collecting UV data for the main peak so we know that the UV data is linear and that the drop off in the ELSD is due to the detector and not due to sample loss to surfaces etc (as I have seen with some molecules).
Going forward we would like to generate a standard curve with the parent compound, transform it to log/log, and then use the curve caluculate the amount of each impuity/related species present. Then use those amounts to calculate the impurity profile. Just like in straight UV, we would be making the assumption that the impurities/RS behave the same as the parent compound.
Does all of this seem to jive with what others have seen when calculating purity by ELSD? Again, I am new to this so any input is appreciated!