Dan » Thu Nov 16, 2006 2:07 pm
danielbor,
The FDA doesn't provide a set calibration frequency, They only require you to set one in an SOP and then follow it.
In general, HPLCs are calibrated every 6 months in the industry. We calibrate every 6 months except for the column heater which we calibrate yearly.
Things to consider:
How old is the system? Older systems may need more frequent calibration.
Do you need to do all of the calibration tests? If you are not going to use the system for gradient work, then don't calibrate it for that but label the system appropriately.
Vendors can give recommendadtions for calibration frequencies as well as guidance for what to calibrate and what tolerances to apply.
Lastly, the USP is developing a new General Information Chapter. You may find it useful. It is USP<1058> Analytical Instrument Qualification. You can find the latest version in the most recent issue of the USP Pharmacopeial Forum (32(6)).
Regards,
Dan