I have an impurity peak that is able to fluorescence. I started be recording the DAD-spectrum of the impurity in UV. Then I set the excitation wavelength to the found UV-max (220, 253, 270 nm), one at the time. Then I made a scan of the emission when the impurity was in the detector. There is a scan function for the excitation wavelength too, but then I had two unknowns (excitation and emission wavelength).
I get emission spectra that look OK, but they all look the same!
Questions:
1: Is it correct to use the UV-max as the excitation wavelength?
2. Why do I get the same emission bands for all excitation wavelengths?
3. I think the noise is quite high (about 0.5 mV at the lowest gain). It is worse S/N then the UV signal. (The lamp is brand new). My idea was to get fantastic QL
Thanks!