Dear Friends

What information we will get from Y-Intercept in method Validation

G.R.Reddy,
AR&D
BIOCON Limited,
India

Quick answer:

Intercept above zero: interferences, blank peaks

Intercept below zero: adsorption of substance

Thank you very much

Is there any articles or publications on this

There can be of course be other reasons for intercept not equal to zero (e.g not linear response). But if the respones is linear the explainations above are quite likely.

Regarding a reference, I think you can find it in a good LC book like Snyder-Kirklands "Introduction to modern liquid chromatography".

Also, you should compute the confidence interval around the intercept. If it includes zero, then you can consider the intercept to be zero for practical purposes.
Mark Tracy
Senior Chemist
Dionex Corp.
5 posts Page 1 of 1

Who is online

In total there are 19 users online :: 1 registered, 0 hidden and 18 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Semrush [Bot] and 18 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry