"The Carrier and make-up flow of the Helium is set manually using a dial at the side of the instrument. The flow of the Hydrogen and the Air gasses is also set this way. It is not a calibrated instrument as used there is no need. The flow is read in Kg/cm2 and the at the moment I found that the following conditions are ok, Carrier Helium: 1.0kg/cm2, Make-up Helium: 0.5kg/cm2, Inj: 130C, Det: 170C, Col: 45C.
I don't have much experience with split flow and the method states 1:50. I looked at the split flow vent and this is manual. There is a glass flute with a tube at the bottom. I have filled it with water and connected it up to the vent to count the number of bubbles traveling out of the vent but none appeared!!!!! So I know it is set at 2-4ml min which is what it is set at. But I don't think that this will make much difference to the resolution and tailing of the peaks. I have no way of measuring the flow of the carrier gas.
OK. Lets sort out your choice of music first. The glass flute is probably a soap film bubble meter. You don't fill it with water, but add a few mls of soap solution to the bulb at the bottom and sweeze the bulb to put a single bubble into the flow from the split vent or septum purge. You then time the bubble to go from the zero to reach the 1, 10, or 100 ml mark. Adding water probably would just block the flow.
You should be able to connect tubing to the flute to get it to measure flows from the detector exit . If you can, then the carrier flow alone from the detector should be about 2 - 3 ml/min, and the flow from the carrier split vent 50 - 150 ml/min.
With luck, setting the detector gases ( air hydrogen, maybe nitrogen makeup pressure to the conditions in the manual will be satisfactory.
But you can check them once you have a soap film meter on the detector outlet. Don't let any soap film or soap solution get into the detector by sweezing the bulb to quickly.
I'm assuming Shimadzu split system is similar to Agilent, in which case if your split vent flow is too low for the head pressure, too much sample will reach the column and your peaks will merge. Overloading would explain your problem.
If noboby from the forum offers detailed instructions for your instrument, and the manual isn't clear, you may want to contact the local Shimadzu technical support on the phone just to clarify which settingsare normal.
Sort out the split ratio, and the music will be much better.
Bruce Hamilton