Dear members of this forum,

I need your help. I have a background line with a relative important noise at 258 nm (the relation of signal - noise) and I think that the problem could be the age of the lamp or any problems in any components of the monochromator.
Thanks in advance for your help,

Diego

Is this noise random? The most common causes or random noise are
1. Not enough light. Can be caused by contaminated flow cell, high absorbance in the mobile phase, mis-alignment of the flow cell
2. Old lamp
3. Bubbles or particles trapped in the flow cell
4. Incorrect settings for the detector
5. Damaged mirror or grating in the detector optics. This is uncommon.

Is there a pattern in the noise?
1. The pump is not stable; the noise will match the cam cycle of the pump
2. A long-period pattern is usually related to room temperature
3. The heat-exchanger on the flow cell is disconnected

Best of luck
Mark Tracy
Senior Chemist
Dionex Corp.
2 posts Page 1 of 1

Who is online

In total there are 2 users online :: 1 registered, 0 hidden and 1 guest (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot] and 1 guest

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry