Discussions about GC-MS, LC-MS, LC-FTIR, and other "coupled" analytical techniques.

5 posts Page 1 of 1
dear friend,
we have validated ms method for isradipine for be studies. my response for LLOQ during validation 45000 but when i performed subject sample anaysis the response drops down to almost half. still my ccs and qcs do not have any problem. all my seniors said that auditors will rise question about it and we have to justify it. we then investigate the problem and it turns out due to the methanol grade. during validation we use brand X and during subject anaysis we use brand Y. now we do not have brand X in stock. can i proceed my analysis provided the justification is due to variation of the brand? did anyone face problem like this? as for me i feel there is nothing wrong but as we are going to face the auditors from outside i just want to know the sugestions from you all.

with regard
pliva
with regard
Donal

How did you prove that it was the methanol? Is this only a suspicion, because this is the only thing that you found to be different between the validation study and the current study? Also, what is the difference between "subject samples" and the validation samples?
How did you prove that it was the methanol? Is this only a suspicion, because this is the only thing that you found to be different between the validation study and the current study? Also, what is the difference between "subject samples" and the validation samples?
yes it is due to methanol. we got one bottle from other lab and checked out with different brand. when i use the brand which we used during validation the response was almost the same but with other brand .
with regard
Donal
How did you prove that it was the methanol? Is this only a suspicion, because this is the only thing that you found to be different between the validation study and the current study? Also, what is the difference between "subject samples" and the validation samples?
yes it is due to methanol. we got one bottle from other lab and checked out with different brand. when i use the brand which we used during validation the response was almost the same but with other brand .
validation sample means the plasma where the drug is spiked externally and subject sample means the blood is is collected from the volunteers during clinical study.
with regard
Donal

If the S/N for LLOQ is still big enough (mostly at least 5), just go ahead.
5 posts Page 1 of 1

Who is online

In total there are 57 users online :: 0 registered, 0 hidden and 57 guests (based on users active over the past 5 minutes)
Most users ever online was 5108 on Wed Nov 05, 2025 8:51 pm

Users browsing this forum: No registered users and 57 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry