Carry over issue of Phosphatidylcholine in Shimadzu LC

Discussions about HPLC, CE, TLC, SFC, and other "liquid phase" separation techniques.

2 posts Page 1 of 1
Hi, I am new to using Shimadzu LC system with Sciex 6500 QTRAP MS. I kept seeing high levels of carry over of different PC molecular species in my analysis despite changing the flush solvents from Methanol to isopropanol to DCM/IPA (10%:90%). Changed a new needle, cleaned or changed injector seal/rotor. Did injection by pass the injector and still seeing the peak in my column. Changed a new column and still there. Different mobile phase mix also doesn't get rid of the peaks.

I was wondering if anyone uses ROMIL solvents and face the same issues before? I usually use OPTIMA LCMS grade solvents from fisher so that is one last thing I can think of trying. Anyone can suggest what can I try or what could be the reason? Thanks.
Anyone can suggest what can I try or what could be the reason?
Phosphatidylcholines are zwitterions, so you must "canceled" it to decrease crossover. You can try to add some soft ion-pairing compounds in the eluent and needle wash solution (0.01 - 0.1 %v): TFA, Trimethylamine or Triethylamine. Start with TFA (note that TFA cause of supression of ionization in ESI). Needle wash solution = MeOH:Chloroform or MeOH:DCM (8:2 to 1:2; must be optimized).
2 posts Page 1 of 1

Who is online

In total there are 12 users online :: 1 registered, 0 hidden and 11 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Google [Bot] and 11 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry