- Posts: 3
- Joined: Fri Aug 11, 2017 2:22 pm
Advertisement
Flame out
Discussions about GC and other "gas phase" separation techniques.
2 posts
Page 1 of 1
Anybody who are using gas chromatography 7890A help me. I am using GC7890A with cp-wax 58 ffap 25m length 0.53mm diameter column. But it is not working tried so many times stilll flame out. Once light out and inject one or two sample then flame out. What i do help mee plzzz
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
1. Try injecting a smaller volume of sample
2. Set FID peak offset from default of 2 to 0.5 I think that's the right way, I'm retired now, so can't run in to check)
3. Check your FID hydrogen flow and air flow, and you shouldn't need make-up gas with 0.5mm i.d. column
2. Set FID peak offset from default of 2 to 0.5 I think that's the right way, I'm retired now, so can't run in to check)
3. Check your FID hydrogen flow and air flow, and you shouldn't need make-up gas with 0.5mm i.d. column
2 posts
Page 1 of 1
Who is online
In total there are 18 users online :: 1 registered, 0 hidden and 17 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 17 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: Bing [Bot] and 17 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science