FID response between 5890 and 6890

Posted: Tue Aug 08, 2017 2:03 pm
by syringe
Does exist a reported equivalence between 5890 FID counts and 6890 FID pA?

Re: FID response between 5890 and 6890

Posted: Thu Aug 10, 2017 3:46 pm
by rb6banjo
If you have both instruments, i would advise that you make standards and see if you can come up with your own empirical relationship. I am unaware of a published relationship. The best practice is to calibrate your new instrument independently - not relative to the old.

Re: FID response between 5890 and 6890

Posted: Thu Aug 10, 2017 4:08 pm
by syringe
Thanks!
The question comes from a friend who do service of GCs but I don't know why he asks for that.

Re: FID response between 5890 and 6890

Posted: Thu Aug 10, 2017 9:45 pm
by Peter Apps
If anyone knows it will be Agilent - get your friend to ask them.

Peter

Re: FID response between 5890 and 6890

Posted: Fri Aug 11, 2017 2:16 am
by Consumer Products Guy
Your signal-to-noise should be similar, that's what's important.

Re: FID response between 5890 and 6890

Posted: Fri Aug 11, 2017 7:09 pm
by James_Ball
Signal counts at baseline will be different depending on gas flows, column bleed, and many other variables.

Also sensitivity will be different with the same variables taken into account.

As for whether the readout on the 5890 was in pAmps, not sure on that one. I don't know if it was a direct current read or if it was converted to different units.