Hi
I am a new user of the HPLC, and I have a new column for the HPLC. The column is an ACCQ.Tag amino acid analyis column. I am trying to do a plate count for this column following the instructions and I don't get a peek. Why??? I run a standard 0.05% Acenaphthene in acetonitrile and let it run 60/40 for 10 min the inject of 10 micro L. I get no peek.
Am I doing something wrong?
Tracy

I don't mean to sound patronizing, but you said you were new to this, so I am checking all the possiblities.

Have you ever seen peaks on this system (with other columns, perhaps)?

Have you double checked to make sure your detector and data system are operating properly (set at the correct wavelength, sensitivity, etc)?

Are you sure your pump is working (mobile phase coming out of the waste line)?

Did you see a small baseline upset marking the dead time? If not, are you sure an injection actually ocurred?
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
2 posts Page 1 of 1

Who is online

In total there are 17 users online :: 2 registered, 0 hidden and 15 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Bing [Bot], Google [Bot] and 15 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry