Hi
I am a new user of the HPLC, and I have a new column for the HPLC. The column is an ACCQ.Tag amino acid analyis column. I am trying to do a plate count for this column following the instructions and I don't get a peek. Why??? I run a standard 0.05% Acenaphthene in acetonitrile and let it run 60/40 for 10 min the inject of 10 micro L. I get no peek.
Am I doing something wrong?
Tracy

My first suspicion would be that you have a problem with the sample dissolution. Make a sample with a much higher concentration, and then dilute.
Also check if the QC method for the column uses indeed 60% acetonitrile and 40% water. Then check if indeed you have made the composition with a higher concentration of acetonitrile, and not by accident the other way round.

Also check your detector/data system settings to confirm the proper wavelength, absorbance range, etc.
-- Tom Jupille
LC Resources / Separation Science Associates
tjupille@lcresources.com
+ 1 (925) 297-5374
3 posts Page 1 of 1

Who is online

In total there are 27 users online :: 2 registered, 0 hidden and 25 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am

Users browsing this forum: Ahrefs [Bot], Semrush [Bot] and 25 guests

Latest Blog Posts from Separation Science

Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.

Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.

Liquid Chromatography

Gas Chromatography

Mass Spectrometry