I am facing a new problem with a brand new column (X-Bridge Shield RP 18 3.5µm 4.6 x 150 mm).
Here are the conditions:
- analyte : micopheniolic acid
- internal standard - naproxen
- matrix: plasma
- volume: 50 µl
- extraction: 3 ml of hexane-diethyether (8:2 v/v) in acid pH (3)
- analysis: LC-MS/MS - API 2000
- mobile-phase: acetonitrile - acetic acid 1mM (8:2 v/v) + 10% isopropanol
- samples are ressuspended with 500 µl of mobile phase and 8 µl are injected.
- flow 0,8 ml/min (split 1:1)
I did a first study with this method using a range of 200 to 30.000 ng/ml and the method works ok for all the samples (2500).
Now I have to quantify the samples in a huge range of 10 to 20.000 ng/ml. I tested different volums to ressuspend the sample and volums of injection, as the follow:
- 200 µl to ressuspend > 25 and 15 µl to inject
- 400 µl to ressuspend > 15 and 10 µl to inject
- 800 µl to ressuspend > 30 and 40 µl to inject
In all cases I have a decreased of 25 - 50% of the analyte response after 50 - 60 samples. If I clean the column with acetonitrile-water (1:1) the response increases again.
So, I know that it is a matrix x column problem, but I do not know how to solve it.
Do I change the mobile phase ??? Maybe try a stronger acid (2 or 5mM) or decrease the organic part (70 or 60%) ?? The ion suppression test was negative.
I hope you can help me.
Thank a lot.