We are currently developing for this method and I have most of it working except where it comes to the final extract composition.
The method says to bring the final extracts up in 10%water/90%methanol. The initial mobile phase is 90%water/10%methanol. When I inject the standards in this solvent I get a spike at the beginning where the solvent front comes through with hits for all the analytes, then when the analytes elute the peak shapes are not great and have low sensitivity. The method uses a solvent divert at the beginning so they have no example chromatograms for this region of the run, since I am not using the divert I do see it.
If I bring my extracts up in 50/50 water methanol this goes away and the peak shapes are excellent. The method states you can't alter any of the extraction steps but you can dilute the samples, but if I dilute to change the composition then I may not be able to achieve the detection limits required. I have never had any luck injecting high levels of methanol into a mostly aqueous mobile phase, is there any trick to making this work, or is the method simply poorly written? Even if I begin with 70%water/30%methanol as the mobile phase I get the same problem with the analytes running through the column with the solvent front.