1 point calibration for chemstation
Discussions about chromatography data systems, LIMS, controllers, computer issues and related topics.
3 posts
Page 1 of 1
I work with a multiple sclerosis group, we are trying to figure out which cannabinoids work best and potency. To help this I purchased an agilent 1100 hplc, I have found walk throughs for everything but how to calibrate. I am using chemstation with a restek standard 1000 ul/ml THCA. Specifically the sample info page (dilition, sample amount etc). Can anyone help? Thanks in advance.
-
- Posts: 88
- Joined: Mon May 16, 2016 11:50 am
you can just search the web for this info, I would do a 5 point curve and aim for >0.999 R2. Just dilute the standard you have
http://www.restek.com/chromatogram/view/LC_GN0553
http://www.restek.com/chromatogram/view/LC_GN0553
-
- Posts: 3594
- Joined: Mon Aug 30, 2004 11:43 pm
After validations, we did 1-point calibrations with Agilent HPLCs, mostly Revision A software. In the calibration table, we entered the active concentration in grams/ml, and set as %Amt or something similar in a different part of Data Analysis, and saved Method.To help this I purchased an agilent 1100 hplc, I have found walk throughs for everything but how to calibrate. I am using chemstation with a restek standard 1000 ul/ml THCA. Specifically the sample info page (dilution, sample amount etc). Can anyone help? Thanks in advance.
Then in Sample Table, the amount of sample was entered in grams. In Dilution, the number of ml of solvent was entered. If there was a subsequent dilution, say 5.0ml diluted to 100ml, then that factor ("20") would go into the multiplier.
To change results to ppm, we'd through an additional 10,000 into the Multiplier column.
So we'd get a single results page, with results in percent, or in ppm; note that when we did as ppm, the Heading of the column would still say "%".
I think I remembered most of it, I'm retired now! I think there's an Agilent PDF on doing this though; the first time I did such on a computer - early 1990s - I had to call HP/Agilent at least 3 times to get it set up, their instruction manual didn't help much.
3 posts
Page 1 of 1
Who is online
In total there are 420 users online :: 0 registered, 0 hidden and 420 guests (based on users active over the past 5 minutes)
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 420 guests
Most users ever online was 4374 on Fri Oct 03, 2025 12:41 am
Users browsing this forum: No registered users and 420 guests
Latest Blog Posts from Separation Science
Separation Science offers free learning from the experts covering methods, applications, webinars, eSeminars, videos, tutorials for users of liquid chromatography, gas chromatography, mass spectrometry, sample preparation and related analytical techniques.
Subscribe to our eNewsletter with daily, weekly or monthly updates: Food & Beverage, Environmental, (Bio)Pharmaceutical, Bioclinical, Liquid Chromatography, Gas Chromatography and Mass Spectrometry.
- Follow us on Twitter: @Sep_Science
- Follow us on Linkedin: Separation Science