currently using a method that runs in 5psi constant pressure mode on a DB-5 column giving me a flow of 5.6ml/min.
I separate heptane and toluene and they come out at 5.2min (heptane) and 8.5min (toluene). never had a problem but now I have changed to a different GC and now the peaks are coming out at 3.9min and 6.9min. the assay is now therefore failing for resolution. remember this is the exact same column and exact same method parameters so it has to be the system right? checked column dimensions and stuff and there fine. the column has never caused issues so far.
now when I check total flow on the system its ~ 18.8ml/min. this cant be right? the column flow is 5.6ml/min and septum purge is ~2ml/min. The total flow on the system I usually use is ~8.5ml/min which seems more in line.
what can cause a significant change in RT when only thing that changes is the system your using?
any help would be great.